Joint Pharmaceutical Analysis Group
|
While ICH validation guidelines provide a framework
for the core activity of measurement validity, actual requirements
are heavily influenced by the purpose and uncertainty of the measurement.
This meeting, reported by Joe Chamberlain, examined key strategies
for the successful validation of analytical methods at all stages
of a medicine's life, from clinical trials through to post-marketing
|
The one-day symposium on method
validation organised by the Joint Pharmaceutical Analysis Group took
place at the Royal Pharmaceutical Society’s London headquarters
on 20 May. Dr Chamberlain is a former editor of The Journal of
Pharmacy and Pharmacology
|
Why robust analytical methods have to exist at every stage of a medicine’s
life
The International Conference on Har-monisation of Technical Requirements
for Registration of Pharmaceuticals for Human Use (ICH) consists of the
regulatory authorities of Europe, Japan and the US and experts from the
pharmaceutical industry in these three regions. The World Health Organization
(WHO), the European Free Trade Area (EFTA), Switzerland and Canada are
also included as observers. The ICH guidelines are invariably referred
to by the principal regulatory authorities for analytical validation,
said Mark Broughton, of Aventis Pharma, Holmes Chapel. Analysts can thus
agree on the core definitions for accuracy (the closeness of agreement
between the true value and that found), precision (the closeness of agreement
between a series of measurements), specificity (verification that the
analytical method responds only to the substance of interest) and range
(the range of sample concentrations over which the analytical procedure
has been shown to have suitable performance). Other useful agreed attributes
include limit of detection, limit of quantitation, linearity and robustness.
The guidelines describe circumstances where revalidation is required
and explain that the degree of revalidation required depends on the nature
of the change. Method maintenance is not covered by ICH but is an expectation
of current good manufacturing practice and should be defined locally
along with a process for carrying out validation and the standards expected. A regulator’s view
Geoff Houghton, of the Medicines and Healthcare products Regulatory
Agency, described the requirements of the EU regulators in new drug applications
or variations, concentrating on the reasons for analysis rather than
the technical aspect. Since July 2003, the product licence (or, in
European terms, the marketing authorisation) has benefited from a
new
common technical document format, with a section devoted to quality
data. Although the bulk of a submission on the analytical method
may describe its technical aspects, it is the validation which is most
critical. Importantly, validation may need to take into account changes
unrelated to the analytical method which may have
an unexpected impact on the validity of an existing method. For example,
a change in formulation from film-coating to enteric- coating may require
identification tests for new excipients, revalidation of existing identity
tests and even revalidation of pharmacokinetic assays used in bioequivalence
studies. Internal procedures should exist so that analytical development
and quality control departments are alerted to such changes.
Paradoxically, the EU licensing authorities require more validation
for a change of method than for the methods designed to support a new
chemical
entity, as the change in method will require cross-validation with
the original method. Dr Houghton stressed that requirements of the
regulatory
authorities often coincide closely with the minimum that competent
analysts would chose to do to ensure that their methods are valid and
robust.
Since the analyst knows more about a particular technique than any
regulator, the analyst should think about why something is being done
rather than
how, and then is more likely to be doing the right thing for satisfying
the authorities, concluded Dr Houghton. Non-chromatographic test methods
Validation of analytical methods is essentially about accuracy and
precision, said Arjen Tinke, of Johnson & Johnson, Beerse, Belgium. For chromatographic
methods, the use of “spiked” samples is often used to evaluate
accuracy but for non-chromatographic methods, orthogonal comparisons
(that is, the application of an alternative assay) are more useful.
For example, loss on drying can be checked by a gas chromatographic
method and laser diffraction for particle size analysis can be checked
by image analysis.
The ultimate test method is one that needs to be applied only once,
and to achieve this with confidence the best precision is required.
To determine
the nature and source of any deviation it is important to develop powerful
analytical test methods. From a speed and cost point of view, it is a
challenge to develop such precise test methods. However, the test methods
that are used in R&D may differ from those that are used in the final
quality control procedure, since in R&D the manufacturing process
and the product may not yet be fully defined.
The use of additional test methods will lead to a more complete mapping
of the product characteristics, or to a better correlation between the
various physical or physiological parameters. Some typical R&D methods
can be used in a later stage, as an orthogonal method in the validation
of the accuracy of a QC method.
Apart from high accuracy and precision, QC test methods should meet additional
requirements with regard to compliance, speed and cost. For the analytical
support of R&D and QC processes, science and compliance should be
considered as equally important. However, for R&D test methods the
cost for analysis may be higher, and the compliance level may be limited. Statistical case studies
Phil Woodward, of Pfizer Global R&D, Sandwich, Kent, presented case
studies where collaborations between analytical chemists and statisticians
added value in pharmaceutical R&D. A typical series of chromatographic
analyses would include interpolated standards which are used to quantify
the analytical samples so bracketed. The required frequency of interpolation
will depend on the amount of drift in the analytical response. Simple
statistical procedures can be used to determine the lowest frequency
consistent with reasonable analyses per day for given degrees of drift,
injection repeatability, and the accuracy required. Thus productivity
can be increased while maintaining data quality.
A high degree of robustness is required in chromatographic methods such
as HPLC which may need to be applied in different laboratories using
different equipment and subject to different degrees of variation. In
the search for the most appropriate operating conditions (column material,
solvents, flow rate, temperature, pH, detector settings) the experimental
design is important. Experiments can be monitored by recording retention
times, resolution, peak responses and tailing factors, after statistical
methods have been applied to give the best coverage of the design space,
detect interactions and ensure increased precision. Overall there is
an increased efficiency in the use of resources, even if a single optimum
is rarely achieved.
Frequent failure of tests for linearity in gas chromatographic assays,
despite a belief that the method was fundamentally sound has led to the
application of statistical methods to resolve this problem.
Mr Woodward noted that simple regression analysis is not useful in testing
for linearity, particularly if the samples are inappropriately spaced
over the test range; naturally tight values close to the origin will
cause a non-zero intercept test to fail rarely, whereas values with a
wide spread at the high end will often trigger a curvature alert. However,
by plotting residuals it can be clearly seen that variation increases
with concentration and hence, with magnitude of response. In this case,
an appropriate modification of the statistical model is to use weighted
least squares. Clinical trials and comparators
An important series of tests often described as pre-formulation is
used to identify a drug candidate’s physical and chemical properties,
said Mark Benger, of Quintiles. These tests are essential in assisting
formulators to make dosage design choices and can help chemists in
development of analytical methods in the early phases of drug development.
The presentation focused on validation of dissolution and stability-indicating
assays, and the implications of using comparators in clinical trials.
Once a set of conditions has been established the analytical method
for dissolution testing should be validated to enable the procedure
to be
deemed fit for purpose. In the early phases it is recognised that a full
validation package is not required to satisfy regulators in the US or
Europe. Clearly, accuracy at the range of expected concentrations should
be validated. “Spiking” experiments, such as those used for
assessing precision and method repeatability, should be included. At
this stage the dissolution procedure would not be fully tested for robustness
but assurances should be made that the drug is stable in the medium of
choice.
With regards to the stability assay, this, too, should be developed and
validated with a view to further modifications to formulations which
are to be expected as the clinical process evolves. The method should
be specific with regards to the main component. The generation of degradants
through limited stress conditions such as heat and light should also
be undertaken.
It is also appropriate that mass balance is evaluated, whereby the sum
of all detected components should approach 100 per cent of the initial
assay value. Often during early development it is not possible to achieve
full mass balance as unknown degradants may have significantly different
ultraviolet absorption compared with that of the parent compound. Other
tests to perform are accuracy, linearity, precision and limits of detection
and quantitation. The early development stability-indicating assay is,
therefore, an important tool in assisting in the understanding of degradation
pathways that may govern the progression of the compound into late-stage
clinical trials.
Another area of interest in the interim stage of a compound’s development
is the use of comparator studies to enable companies to test their candidate
product against the market leader in blind clinical trials. Once the
marketed product is blinded, for example by over-encapsulation, it is
considered to be a new product which requires its own analytical validation.
There are requirements to check that the use of a particular blinding
process has not altered the release rate of the product or indeed the
chemical stability. Late stage pharmaceutical development
Kevin Ryan, of Pfizer Analytical, Sandwich, Kent, provided a useful
definition of validation. It is the process by which assurance is provided
that
we have developed a method that is fit for its intended purpose, ie,
we understand clearly what the performance requirements are and we
are confident that the method can meet these routinely.
Validation data are used to judge the capability of a procedure to
generate analytical results of the expected quality, reliability and
consistency.
It reflects the method development process and consequently poor development
inevitably means a poor validation data set. There is considerable guidance
for method validation for the industrial analyst. Vital documents are
ICH Q2a and Q2b, although these should not be applied blindly but on
a case-by-case basis. Important documents include the US Food and Drug
Administration “Draft guidance for industry — analytical
procedures and method validation, and the “FDA reviewer guidance — validation
of chromatographic methods”. Measurement uncertainty
Analysts generally express the quality of their measurements in terms
of a measurement error and are frequently over-optimistic in their
estimations, often ignoring significant sources of error. However,
measurement error is not a useful parameter for estimating and expressing
measurement quality, said Steve Wood, from the Laboratory of the
Government Chemist. Measurement uncertainty, which is estimated from
what is known
about a measurement rather than by speculating about unknown residual
errors, is a more robust methodology and should be used instead.
Measurement uncertainty is an effective way of expressing the errors
that are inherent
in any measurement method. It is a concept used increasingly by analytical
scientists and is being adopted in international standards for certification
and accreditation of measurement laboratories.
Evaluation of the uncertainty of any measurement method should be
an integral part of method validation and knowledge of the uncertainty
of
a result is essential for comparison against specification or regulatory
limits. The concept of uncertainty itself is straightforward but implementing
it often causes confusion.
There are four key steps involved in quantifying uncertainty: what
needs to be measured should be specified in detail; for each stage
of the measurement
procedure sources of uncertainty should be listed, identifying what
causes the result to change; the uncertainty components then must be
quantified;
and, finally, the values are combined to give a quantitative measure
of uncertainty as a range. New technologies
No clear guidelines exist for the validation criteria required for
a new technology and consequently the process for the introduction of
something novel can be somewhat complex, said Manuel Sanchez-Felix,
of Lilly, US. Indeed, although the FDA states its support for the
development
and use of the most appropriate instrumentation, it has also recognised
that the use of rare or exotic systems places a burden on the regulatory
laboratory, and may delay registration. Until recently this potential
to delay registration may have deterred the industry from introducing
new technology but there has been a welcome change in approach that
encourages the use of new technologies.
Vehicles for the introduction of new technologies include user groups
(for example, the Pharmaceutical Analytical Science Group, the American
Association of Pharmaceutical Sciences, and the Joint Pharmaceutical
Analysis Group), working groups (particularly the New Technology Forum,
a UK partnership established by the Royal Pharmaceutical Society, the
Medicines and Healthcare products Regulatory Agency and the pharmaceutical
industry), and the regulatory authorities.
Now that the industry has incentives and rapidly evolving processes,
the regulatory environment is changing to one of process understanding
and the guidelines now being developed by the regulatory authorities
act as catalysts for the introduction of new technology, said Dr Sanchez-Felix. |