Joint Pharmaceutical Analysis Group
|
Practising pharmaceutical analysts specialising
in near-infrared spectroscopy gathered with regulators and teachers
to examine and discuss the current state of the art. Joseph Chamberlain
reports
|
This one-day symposium organised by the Joint
Pharmaceutical Analysis Group took place on 13 October at the Royal Pharmaceutical
Society’s London headquarters
|
Developments in NIR spectrometry in relation to pharmaceutical analysis
Near-infrared spectroscopy (NIRS) is routinely used in the pharmaceutical
industry for applications such as raw material identification and moisture
determination. Although these tests can be carried out rapidly using
NIRS the major advantages of the technique are not being fully exploited,
said Mark Smith, of Pfizer, Sandwich, Kent. For example, moisture content
is rightly regarded as a critical parameter for release of a product,
yet it is usually measured after compression, when nothing can be done
if the specification fails.
The new focus is to develop NIR methodology for in-line and on-line
purposes, requiring a radically different approach. Although the classical
headings
of method validation remain relevant for NIRS methods, some of
the tests and outcomes can be construed
differently.
For many applications a different approach may be justified based on
scientific judgement, provided the final method is fit for its intended
purpose. Dr Smith defined reproducibility as the ability to transfer
a method created on a single instrument to a second instrument and
this is still regarded as one of the big challenges in modern NIRS.
Barriers
to successful transference include hardware differences and a lack
of understanding. The presentation showed some specific examples of
calibration
transfer in the pharmaceutical industry and methods for achieving acceptable
transfer.
NIRS can be used to monitor the fluid bed dryer for a predetermined
water content, hence ensuring the material is dry, but not over-dry,
before
continuing the process. This approach also optimises time used by the
drying equipment. Similarly, off-line content uniformity performed
after compression on cores or coated tablets, while useful as a quality
control
device, has no added value in the manufacturing process. NIRS can be
used to monitor the blending stage so the process can be terminated
when a satisfactory blend is achieved and, again, no additional plant
time
is wasted.
There is much to be learnt by attention to the relevant parameters
in the development of the NIRS method used for specific purposes, concluded
Dr Smith. Application in a solid-dosage facility
Thorsten Herkert described how NIRS is used as the only analytical
control method in the AstraZeneca solid dosage facility in Plankstadt,
Germany.
A three-storey building houses all the elements of manufacture from
reception of raw materials to the finished tablets. NIRS is used
at five different
stages of the production and there is an integrated IT system covering
all applications.
Brimrose NIR-analysers with acousto-optic tuneable filters are used
for all applications. The process is fast and there are no moving
parts. The warehouse instrument uses a standard analyser. The dispensaries
and
the fluid bed dryer stations use a multiplexed analyser with probes,
and the blending monitor is a novel, purpose-built analyser measuring
through a sapphire window. The compression monitor is also a novel,
purpose-built analyser for non-destructive tablet measurement.
Each analytical station is designed for a particular application,
but has a consistent operator interface with simple control panels
linked
to bespoke software. At the dispensary, the operator scans the material
barcode, the analyser is enabled for operation and the operator is
prompted through the process. Spectra are collected with the in situ
probe, identification
is confirmed and reported, with all results and metadata sent to
a data storage system.
At the blender, there is real-time monitoring for end-point control.
A stop signal is transmitted to a blender control unit on reaching
the end point, when the analyser is returned to the docking station
and data
are downloaded to the data storage system. The tablet analyser samples
during the compression run, with analysis of single tablets for identity,
active agent content and tablet
hardness.
The concept of total quality management offers real-time process
monitoring and control of the relevant stages of the tablet manufacturing
process.
Additionally, it uses NIRS technology to provide improved process
understanding, which, in turn, gives enhanced assurance of finished
product quality
and a solid platform for real-time release and a right-first-time
philosophy. Less obvious benefits include reduced or no conventional
end-product
testing, reduced lead times, and reduced necessity for large stocks. Future generations of NIRS instruments
NIRS has been developed for at-line, on-line and in-line applications,
the ultimate aim being true quantitative analysis of in situ chemical
content in process vessels in real-time. The main limitations of NIRS
still remain, with problems of quantitative assessment when sample
parameters such as size, density, and particle distribution vary, said
Jonas Johansson, of AstraZeneca R&D, Mölndal, Sweden. The
standard rationale to deal with these problems is to apply multivariate
modelling, but this can only be applied to a limited extent. When sample
parameters lie outside the calibration, NIRS cannot be used for reliable
analysis.
Process analytical technology is a prime area where spectroscopic tools
are appropriate. NIRS fulfils the requirements of high capacity, accuracy,
selectivity and robustness, allowing in situ measurements. It is rich
on information, and has the potential to yield an in-depth understanding
of the process being monitored.
The challenge for future instrumentation lies in the fact that useful
information is overlaid on top of a huge amount of background information
such as excipient absorbance, light-scattering effects and detector variance.
The difficulties lie in calibration transfer, the structure of the sample
and global calibrations.
The main optical properties of tablets that can be utilised are absorption
for chemical content and edge effects, and scattering for the physical
parameters of the sample.
A recent development is the use of time-resolved spectroscopy for the
analysis of tablets; photon migration in turbid samples can be calculated
and both structural and chemical information can be extracted. By extending
conventional NIRS into the time domain, a better correlation to chemical
content can be attained for samples of varying physical properties.
Gas in scattering media absorption spectroscopy (GASMAS) is a technique
that measures the amount of free gas dispersed in solids and thus can
be applied to monitoring porosity or disintegration of tablets. When
batches were produced with different compression forces, the tablet hardness
and
GASMAS signal were well correlated. Similarly, sieved batches of granulate
showed a correlation between granule size and the GASMAS signal.
Future challenges will include the development of more selective tools
and more robust quantitative tools, concluded Dr Johansson. Process understanding to control
In everyday life, visualisation is a good means of identifying a problem — you
can see something has gone wrong, said Fiona Clarke, of Pfizer, Sandwich,
Kent. Deficiencies in a pharmaceutical product are not so obvious and
chemical methods are needed. NIRS is an ideal method because it can be
manipulated to produce detailed images of pharmaceutical raw materials
and finished products. At Pfizer, NIRS is used at the laboratory scale
to study and understand pharmaceutical processes and to use this understanding
to investigate and solve problems at the manufacturing level. It is the
basis of an imaging technique for raw materials and finished products.
This is done by generating a component image, then defining contours
based on pixel intensity.
In the field of process development, the technique can be used to study
the changes in how components form clusters on compression, the variation
in micromixing at various scales, and the effect of using active compound
of differing particle size on the distribution of the compound in the
tablet matrix.
In a study of dissolution, NIRS was used to produce images of a tablet
and to derive from these images the distribution of the components, including
a measure of the particle sizes of individual components within the mixture.
It was possible to correlate such properties with dissolution rates to
the extent that, in model systems, the dissolution rate of a tablet could
be predicted from inspection of the image.
NIR microscopy has rapidly evolved over the past six years. In research
it can be used to build process knowledge and understanding. Although
still in its infancy in manufacturing, the concept is well understood
and advances in this field are expected, concluded Ms Clarke. Counterfeiting
Near-infrared spectroscopy is an excellent method of analysis for counterfeit
medicines, said Tony Moffat, of the School of Pharmacy, University
of London. The school has developed an NIRS procedure for the Korean
Food and Drug Administration in Busan as an alternative to the time-consuming
high-performance liquid chromatography method for the identification
of counterfeit impotency drugs.
Counterfeit Cialis tablets investigated could be easily differentiated
from authentic tablets just by examining their NIR spectra. Further
analysis by HPLC showed them to contain sildenafil (the active ingredient
of Viagra)
and not tadalafil. Counterfeit Levitra tablets could similarly be differentiated
from authentic Levitra tablets by the use of NIRS. One batch contained
sildenafil and tadalafil instead of vardenafil. Counterfeit Viagra
tablets were differentiated from authentic tablets by NIRS, but with
more difficulty
because they did contain the correct ingredient (sildenafil). Apart
from comparing the whole tablet to see if it has been manufactured
by the
claimed pharmaceutical manufacturing company, NIRS can also be used
to measure the amount of active drug in the tablet to ensure its
quality,
and can even discriminate the site of manufacture of the medicine.
In recent years, counterfeit medicinal products have been suspected
to have entered the legal pharmaceutical supply chain, said Andy Charvill,
of the Medicines and Healthcare products Regulatory Agency. The agency
has used NIRS for screening suspicious samples and comparing spectra
with those obtained from authentic product provided by the holder of
the marketing authorisation. NIRS provides data that will confirm authenticity
as long as the full set of standards is available. However, witness
statements
still need to include confirmatory analyses such as HPLC. Although
NIRS has the undoubted advantages of speed and non-destructiveness,
there
is still this requirement for comparators, which includes multiple
batches covering different sites of manufacture and covering all sources
of excipients
and manufacturing processes.
Counterfeit Cialis, Reductil and Lipitor have so far been confirmed
within the UK legitimate supply chain. In each of these cases NIRS
was used
to demonstrate that the suspicious samples differed sufficiently from
the authentic material to require further analysis and to alert the
enforcement group that further investigation was required. For Cialis
the spectra
obtained showed a wide variation between samples, in both absorbance
and wavelength, indicating that the samples provided came from a number
of different formulations and manufacturing batches or sources. The regulatory perspective
NIRS is one of the most frequently used technologies in process analysis,
being used for blend uniformity, monitoring of drying processes, moisture
content, suspension homogeneity and potency. However, treatment of
NIRS as a methodology is no different from any other analytical method,
said Gary Ritchie, of the US Pharmacopeia (USP). The USP project team
on process analytical technology (PAT) was started by the USP to make
PAT relevant to the pharmaceutical industry and support the PAT framework
started by the Food and Drug Administration. The project team has a
broad spread of industry representatives, to recommend revisions and
additions of general monographs to facilitate the use of PAT within
the industry. One of these monographs, “Near infrared spectrophotometry”,
has been rewritten extensively and is being reviewed by the general
chapter expert committee.
Abigail Moran, of the Medicines and Healthcare products Regulatory
Agency also discussed the hot topic of PAT in her presentation on the
regulator’s
perspective on NIRS. She reported that as yet no PAT applications have
been received by the agency; however, an EU PAT team is established and
has been in dialogue with industry to ensure that regulators are ready
for such submissions and many companies are gearing up for PAT. Applicants
wishing to use NIRS tests in support of a marketing authorisation should
consult the guidance note from the Committee for Proprietary Medicinal
Products on the use of NIRS by the pharmaceutical industry and the data
requirements for new submissions and variations. Other general sources
of guidance would be found in the European Pharmacopoeia general monograph
on NIRS and the previously mentioned USP general monograph. The Pharmaceutical
Analytical Sciences Group had also issued guidelines for the development
and validation of NIRS.
Dr Moran noted that feedback from the pharmaceutical assessors suggest
that the majority of applications are through post-licensing and most
are qualitative tests for identification, although the number of quantitative
tests is increasing. Common deficiencies in
applications include the absence of instrument description and minimal
description of chemometric methods. GMP inspectors see NIRS as an up-and-coming
technology, being used in a diversity of ways including raw material
identification, blend homogeneity, optimisation of excipient profile,
particle sizing, water content, drug content, control of coating processes,
leak testing, and packaging identification. Thus, concluded Dr Moran,
the agency expected increasing use of NIRS methods to support applications,
and greater involvement in process analytical technology, particularly
including more dialogue with the industry. |